Why does a pH electrode respond more slowly in cold solutions?
Answer: A
Glass membrane resistance roughly triples for each ten degree fall, so the time constant of the measuring circuit lengthens and readings take much longer to settle, sometimes minutes near zero degrees. Low temperature electrodes use a lower resistance glass. Buffer values do change but that is a calibration matter, crystallisation blocks the junction rather than slowing the glass, and the Nernst slope falls with temperature but never reverses. Patience, or a warmer sample, is the practical answer.