Machine-translated by Google. Questions are authored in English.
Top

Discussion

A digital voltmeter reading of 0.000 with the leads open and 0.002 with them shorted on the two volt range indicates

  • A.that the lead resistance is two milliohms
  • B.a small offset of two counts
  • C.a fault in the analogue to digital converter of the instrument
  • D.that the auto-zero circuit is working correctly as designed

Answer: B

Two counts on that range is two millivolts, and it appears only when the input is completed, so it is a thermal or contact offset in the leads rather than in the instrument. It matters when measuring millivolt signals and is dealt with by noting and subtracting it.

No comment is present. Be the first to comment.
Loading…

Post your comment