Checking an instrument's zero after a set of readings as well as before is worth doing because

  • A.a drift during the readings would otherwise go unnoticed
  • B.the zero cannot be checked before readings are taken
  • C.the second check improves the resolution of the instrument
  • D.standards require every instrument to be zeroed twice

Answer: A

Correct answer: a drift during the readings would otherwise go unnoticed

Checking the zero at both ends brackets the readings: if it has moved, the readings taken between are suspect and can be corrected or repeated. Checking only beforehand assumes stability over the whole session, which for a sensitive instrument warming up or in a changing ambient is not a safe assumption.