What is meant by the slope of a pH electrode?

  • A.The rate at which its reading drifts with time
  • B.The angle at which it must be mounted in the vessel
  • C.The gradient of the calibration buffer against temperature
  • D.Its millivolt change per unit of pH

Answer: D

Slope is the measured millivolts per pH unit, expressed as a percentage of the theoretical 59.16. A new electrode gives 98 to 100 per cent; below about 92 per cent it should be replaced, because the response has become compressed and readings far from the calibration points will be wrong. Drift with time is a separate symptom, usually of a fouled junction, mounting angle matters only for the electrolyte level, and buffer temperature behaviour is tabulated separately. Slope and offset together are what a two point calibration determines.