Spectrophotometry and Colorimetry Questions and Answers

21.

Colour measurement instruments typically report reflectance at intervals of

Answer: A

Ten and twenty nanometre intervals are both common in industrial instruments. Finer intervals bring little benefit for smooth reflectance curves. Ten and twenty nanometre intervals are both common. Either suffices for a smooth curve.

Please log in or register to report this question.

22.

Assertion (A): A ten nanometre interval is adequate for most colourant reflectance curves.
Reason (R): Such curves are smooth and contain no narrow features to be missed.

Answer: B

Correct answer: Both A and R are true, and R is the correct explanation of A

Both are true and the reason gives the justification. A sampling interval must be fine enough to follow the fastest change in the curve, and the reflectance curves of ordinary colourants are broad and smooth, so ten nanometres captures them. A source with narrow emission lines is a different matter and needs a much finer interval.

Please log in or register to report this question.

23.

Consider the following statements about wavelength interval.
1. Smooth reflectance curves are adequately sampled at ten nanometres.
2. Narrow band emission requires a much finer interval.
3. A finer interval always improves the measured colour values.

Answer: C

Correct answer: 1 and 2 only

Statements 1 and 2 are correct: smooth curves are adequately sampled at ten nanometres and narrow band emission is not. Statement 3 is wrong - once the interval is fine enough to follow the curve, reducing it further adds nothing but noise and measurement time, since there is no structure left to resolve.

Please log in or register to report this question.

Answer: D

Instruments measuring fluorescent samples extend further into the ultraviolet. The visible range is the minimum required for tristimulus calculation. Fluorescent work extends into the ultraviolet. The visible alone suffices for ordinary samples.

Please log in or register to report this question.

Answer: A

Holmium or didymium filters provide sharply defined absorption bands. A wavelength error shifts every reading and corrupts the computed values. Holmium and didymium filters are used for the check. Their absorption bands are sharply defined.

Please log in or register to report this question.

26.

Assertion (A): A wavelength scale error affects every measurement the instrument makes.
Reason (R): Reflectance is attributed to the wrong wavelength throughout the spectrum.

Answer: B

Correct answer: Both A and R are true, and R is the correct explanation of A

Both are true and the reason gives the mechanism. If the wavelength scale is displaced, every reflectance value is recorded against the wrong wavelength, so the whole curve is shifted and every tristimulus value computed from it is wrong. The error is systematic and appears in every measurement, which is why the wavelength scale is checked with a filter having sharp absorption bands.

Please log in or register to report this question.

Answer: C

Correct answer: the correctness of the reflectance scale it reports

Photometric accuracy concerns the vertical axis: whether a surface of 40 per cent reflectance is reported as 40 per cent. It is checked with grey tiles of certified reflectance spanning the range. Wavelength accuracy concerns the horizontal axis, and repeatability is a separate property again - an instrument can be repeatable and inaccurate at once.

Please log in or register to report this question.

28.

Match each instrument performance characteristic with its check.
List I
A. Photometric scale
B. Wavelength scale
C. Zero and stray light
D. Repeatability
List II
1. a black trap at the measuring port
2. grey tiles of certified reflectance
3. repeated readings on a stable tile
4. a filter with sharp absorption bands

Answer: D

Correct answer: A-2, B-4, C-1, D-3

Each characteristic has its own check. The photometric scale is verified with grey tiles of certified reflectance, spanning the range rather than only the white end. The wavelength scale is checked with a filter having sharp absorption bands at known positions. Zero and stray light are checked with a black trap. And repeatability is measured by repeated readings on one stable tile.

Please log in or register to report this question.

Answer: A

It is quoted as a mean colour difference from a set of reference tiles. It limits how well measurements can be shared between sites. It is quoted as a mean difference over reference tiles. Specifications state it explicitly.

Please log in or register to report this question.

30.

Assertion (A): Colour data shared between two sites should ideally come from profiled instruments.
Reason (R): Inter-instrument differences would otherwise be mistaken for real colour differences.

Answer: B

Both are true and the reason gives the risk, and the difference can exceed the tolerance itself. Profiling to a master instrument is the usual remedy. Differences can exceed the tolerance itself. Profiling reduces them considerably.

Please log in or register to report this question.

Loading…
Tags: Spectrophotometry and Colorimetry Questions and Answers || Spectrophotometry and Colorimetry MCQ Questions and Answers || Spectrophotometry and Colorimetry GK Questions and Answers || Spectrophotometry and Colorimetry GK MCQ Questions || Spectrophotometry and Colorimetry Multiple Choice Questions and Answers || Spectrophotometry and Colorimetry GK || GK on Spectrophotometry and Colorimetry || Colour Science and Measurement Questions and Answers || Colour Science and Measurement MCQ Questions and Answers || Colour Science and Measurement GK Questions and Answers || GK on Colour Science and Measurement