Analog Ammeters and Voltmeters Questions and Answers

Answer: A

Correct answer: the deflection settles at a definite value for a given input

Without a controlling torque the pointer would swing to the end of the scale for any input at all, since nothing would oppose the deflecting torque. The controlling torque grows with deflection, so the pointer stops where the two balance - and it also returns the pointer to zero when the input is removed.

Please log in or register to report this question.

Answer: B

It depends on the velocity of the moving system, so it opposes the motion and dies away as the pointer settles, without shifting the final reading. Too little gives oscillation and too much gives a sluggish instrument.

Please log in or register to report this question.

Answer: C

The former moving in the strong field of the permanent magnet has currents induced in it whose reaction opposes the motion, which is both effective and free of any moving contact. Air vane damping is used where no such field is available.

Please log in or register to report this question.

Answer: D

Correct answer: its field is too weak and would be distorted by a damping disc

Eddy current damping needs a strong magnetic field for the disc to move in, and the moving iron instrument’s own field is weak. Introducing a permanent magnet for damping would distort that field and upset the reading, so a light vane moving in a sealed chamber is used instead.

Please log in or register to report this question.

Answer: A

A weight on an arm gives a controlling torque proportional to the sine of the deflection, which needs the instrument level and gives a cramped scale at the lower end. Springs work in any position, which is why portable instruments use them.

Please log in or register to report this question.

Answer: B

Spring torque grows in proportion to the deflection, so a deflecting torque proportional to the measured quantity gives equal divisions. A square-law deflecting torque gives the cramped scale of a moving iron instrument.

Please log in or register to report this question.

Answer: C

The torque follows the instantaneous current and the inertia of the moving system averages it, so on an alternating current of zero mean the pointer does not move at all. That is why such an instrument is a direct current instrument only.

Please log in or register to report this question.

Answer: D

Correct answer: proportional to the current

With a permanent magnet supplying a constant field, the torque on the coil is proportional to the current in it, and the spring torque is proportional to the deflection. The two balance at a deflection proportional to the current, which gives an evenly divided scale - one of the movement’s principal advantages.

Please log in or register to report this question.

Answer: A

Averaging a sinusoid over a cycle gives nothing, so the pointer sits at zero while quivering slightly. Adding a rectifier makes the movement usable on alternating current at the cost of a waveform-dependent calibration.

Please log in or register to report this question.

Answer: B

Correct answer: ohms per volt

\[ \mathrm{Sensitivity} = \frac{1}{I_{fsd}}, \] expressed in ohms per volt. It states how much resistance the instrument presents for each volt of range, so a higher figure means less current drawn from the circuit and less disturbance to what is being measured.

Please log in or register to report this question.

Loading…
Tags: Analog Ammeters and Voltmeters Questions and Answers || Analog Ammeters and Voltmeters MCQ Questions and Answers || Analog Ammeters and Voltmeters GK Questions and Answers || Analog Ammeters and Voltmeters GK MCQ Questions || Analog Ammeters and Voltmeters Multiple Choice Questions and Answers || Analog Ammeters and Voltmeters GK || GK on Analog Ammeters and Voltmeters || Electrical Measurements Questions and Answers || Electrical Measurements MCQ Questions and Answers || Electrical Measurements GK Questions and Answers || GK on Electrical Measurements