Measurement Errors and Standards Questions and Answers

Answer: A

Correct answer: in quadrature, as the root of the sum of squares

\[ u = \sqrt{u_1^2 + u_2^2} \] for independent uncertainties. Simple addition would assume both errors take their maximum value in the same direction at once, which is improbable, and would overstate the combined uncertainty. Quadrature is the correct treatment when the two are genuinely independent.

Please log in or register to report this question.

Answer: B

Differentiating the product shows that the fractional uncertainties are what carry through, so a one per cent uncertainty in each gives about one and a half per cent in the product. Mixing absolute and relative quantities is the usual error.

Please log in or register to report this question.

23.

A resistance is found from a voltage known to 1 per cent and a current known to 2 per cent. The resistance is uncertain by about

Answer: C

Correct answer: 2.2 per cent

For a quotient the relative uncertainties combine in quadrature: \[ \sqrt{1^2 + 2^2} = \sqrt{5} = 2.24\ \mathrm{per\ cent}. \] Note how the larger contribution dominates - improving the 1 per cent measurement to 0.5 per cent would change the total only to 2.06 per cent. Effort belongs on the worst term.

Please log in or register to report this question.

Answer: D

Correct answer: realises the unit from its definition

A primary standard embodies the unit directly, by an experiment traceable to the definition rather than by comparison with anything else - a Josephson array for the volt, a quantum Hall device for the ohm. It is held at a national laboratory and is not used for routine work.

Please log in or register to report this question.

Answer: A

Correct answer: is calibrated against a primary standard and held in a laboratory

The secondary standard is the intermediate link: it is compared periodically against a primary standard and then serves as the reference within its own laboratory. The hierarchy exists so that the primary standard is disturbed as little as possible while accuracy still propagates down the chain.

Please log in or register to report this question.

Answer: B

Correct answer: is used for the routine calibration of instruments

The working standard is the bottom of the chain and the one that does the daily work, checked against a secondary standard at intervals. Because it is in constant use it is expected to drift, which is why the interval between its own calibrations is a matter of documented policy.

Please log in or register to report this question.

Answer: C

Correct answer: an unbroken chain of comparisons back to a national standard

Traceability means every link in the chain is documented, each with its own stated uncertainty, so the measurement can be related to the national realisation of the unit. Buying from a reputable supplier is not traceability; a certificate naming the standards used and their uncertainties is.

Please log in or register to report this question.

Answer: D

Correct answer: the corrections found and the uncertainty of the calibration

A certificate is a measurement result, not a verdict. It records what the instrument read against the standard at each point, the correction implied, and the uncertainty of the calibration itself - because a calibration is a measurement and carries an uncertainty like any other. A bare pass or fail conveys none of that.

Please log in or register to report this question.

Answer: A

An instrument that has proved stable over several cycles can have its interval extended, and one whose reading matters greatly is recalibrated more often. Fixing every interval at a year regardless is convention rather than reasoning.

Please log in or register to report this question.

Answer: B

The error was presumably present for some part of the interval, so anything the instrument was used to accept or reject has to be reconsidered. That obligation is why calibration records are kept with the test results.

Please log in or register to report this question.

Loading…
Tags: Measurement Errors and Standards Questions and Answers || Measurement Errors and Standards MCQ Questions and Answers || Measurement Errors and Standards GK Questions and Answers || Measurement Errors and Standards GK MCQ Questions || Measurement Errors and Standards Multiple Choice Questions and Answers || Measurement Errors and Standards GK || GK on Measurement Errors and Standards || Electrical Measurements Questions and Answers || Electrical Measurements MCQ Questions and Answers || Electrical Measurements GK Questions and Answers || GK on Electrical Measurements