Transducers in Measurement Questions and Answers

81.

A resistance wire strain gauge with a gauge factor of 2 is bounded to a steel structural member subjected to a stress of 100 MN/m2. The modulus of elasticity is 200 GN/m2. Percentage change in the value of gauge resistance due to the applied stress is __________.

Exam Year(s)

RSMSSB JE - 2025, RSMSSB - 2025,

Answer: C

The resistance changes by 0.1%.

First find the strain from Hooke's law, then multiply by the gauge factor, \(\frac{\Delta R}{R} = G\varepsilon\).

\[ \begin{aligned} \varepsilon &= \frac{\sigma}{E} = \frac{100 \times 10^{6}}{200 \times 10^{9}} = 5 \times 10^{-4} \\ \frac{\Delta R}{R} &= 2 \times 5 \times 10^{-4} = 10^{-3} \end{aligned} \]

A fractional change of \(10^{-3}\) is 0.1%. 0.01% and 1% come from slipping a factor of ten in the prefixes M and G, and 10% would mean an impossible strain for steel.

Please log in or register to report this question.

82.

An accelerometer has a seismic mass of 0.05 kg and spring constant of 3 × 103 N/m. If the maximum mass displacement is ±0.02 m, the maximum measurable acceleration before the mass hits the stop is :
[Given : √2 = 1.4145, √3 = 1.732]

Exam Year(s)

RSMSSB JE - 2025, RSMSSB - 2025,

Answer: B

The maximum measurable acceleration is 1200 m/s2.

The mass hits the stop when the spring is deflected by 0.02 m. At that point the spring force equals the force needed to accelerate the seismic mass:

\[ \begin{aligned} kx &= ma \\ a &= \frac{kx}{m} = \frac{3 \times 10^{3} \times 0.02}{0.05} \\ &= \frac{60}{0.05} = 1200\ \text{m/s}^{2} \end{aligned} \]

The square-root values printed with the question are not needed. 600 and 300 come from halving or quartering the displacement, and 4.9 m/s2 is about half of g, which has nothing to do with the data.

Please log in or register to report this question.

83.

A thermistor is suitable for :

Exam Year(s)

MH SET - 2024, SET Exam - 2024,

Answer: C

A thermistor is suitable for low span but high sensitivity.

A thermistor's resistance changes by several percent per degree, much more than a metal resistance thermometer's, so it is very sensitive. But its resistance-temperature relation is strongly non-linear (exponential), and its usable range is limited to a few hundred degrees, so its span is small.

High span with low sensitivity describes thermocouples or platinum resistance thermometers better. A device with both high span and high sensitivity, or with both low, does not describe the thermistor's characteristic trade-off.

Please log in or register to report this question.

84.

Match the following :

List – IList – II
a. 3½ digit1. Guarding/shielding
b. Pt – 1002. Bridge
c. RTD3. Temperature
d. Interference4. DVM

Codes : a b c d

Exam Year(s)

UGC NET - 2012,

Answer: B

A 3 and a half digit display is the format of a digital voltmeter, so a matches 4. Pt-100 is a platinum resistance sensor that measures temperature, so b matches 3. An RTD is read through a bridge circuit, so c matches 2. Interference is reduced by guarding and shielding, so d matches 1. The code is 4 3 2 1. The code 1 2 3 4 is the reverse, and 3 1 4 2 and 3 4 1 2 mismatch the first pair. The answer is 4 3 2 1.

Please log in or register to report this question.

Loading…
Tags: Transducers in Measurement Questions and Answers || Transducers in Measurement MCQ Questions and Answers || Transducers in Measurement GK Questions and Answers || Transducers in Measurement GK MCQ Questions || Transducers in Measurement Multiple Choice Questions and Answers || Transducers in Measurement GK || GK on Transducers in Measurement || Electrical Measurements Questions and Answers || Electrical Measurements MCQ Questions and Answers || Electrical Measurements GK Questions and Answers || GK on Electrical Measurements