Spectrophotometry Fundamentals Questions and Answers

81.

Which of the following would not shift a calibration line's intercept?

A change in the analyte's molar absorptivity

A blank that omits one of the reagents

Cells inserted the wrong way round

An unaccounted absorbing impurity in the solvent

Answer: A

A change in molar absorptivity alters the slope rather than the intercept, since it multiplies the concentration term. The other three all add a constant absorbance present even at zero analyte. Whether a defect changes a line's slope or its intercept is the first distinction to draw when a calibration looks wrong, and it halves the list of possible causes immediately.

Please log in or register to report this question.

Answer: B

The reference path carries a matched cell containing the blank, so solvent and cell absorption cancel along with any source or detector drift. An empty path would leave the cell and solvent uncorrected, a second sample would cancel the analyte too, and a fixed filter corrects nothing that varies. Arranging the reference to differ from the sample in exactly one respect is what makes the comparison meaningful.

Please log in or register to report this question.

Answer: C

A rotating mirrored disc directs the beam first through the sample and then through the reference, many times a second, so one detector sees both and their ratio is formed electronically. Ambient rejection is a useful side effect of the modulation rather than the purpose, and neither slit nor detector is switched. Using one detector for both paths removes any mismatch between two detectors, which would otherwise reappear as exactly the error the arrangement exists to remove.

Please log in or register to report this question.

Answer: D

Any difference in sensitivity or drift between two detectors would appear in the ratio as an error indistinguishable from absorbance, which is precisely what the double-beam arrangement is trying to eliminate. Speed and amplification are secondary, and cell matching is needed regardless. Removing a component from a comparison, rather than trying to match two of them, is nearly always the more robust choice, and the chopper is what makes it possible.

Please log in or register to report this question.

Answer: A

A split-beam design divides the light permanently and gives each path its own detector, so both are measured at once with no moving parts. The price is that the two detectors must be matched and must stay matched. Chopping trades mechanical complexity for freedom from detector mismatch, and splitting trades the reverse - which is a clean illustration of two designs solving the same problem by moving the difficulty to different places.

Please log in or register to report this question.

Answer: B

Without a matched cell in the reference path the sample cell's walls and the solvent would appear as absorbance, since only one path would contain them. Beam intensity, detector balance and chopper operation are not the reason. The double-beam arrangement cancels whatever is common to the two paths, so anything that must cancel has to be placed in both.

Please log in or register to report this question.

87.

A double-beam instrument reads \( 0.65 \) with the sample in place. If the blank alone reads \( 0.05 \) against air, the sample's corrected absorbance is

Answer: C

Subtracting the blank's own contribution:

\[ A = 0.65 - 0.05 = 0.60 \]

giving 0.60. In normal use the blank sits in the reference path and the subtraction happens automatically, which is exactly why the instrument reads 0.60 directly when set up properly. Understanding what the automatic correction is doing matters when a reading has to be checked or a blank is suspect.

Please log in or register to report this question.

Answer: D

Absorbance is the logarithm of a ratio, and the reference supplies the denominator, so a different reference gives a different number for the same sample. Detector gain, wavelength calibration and lamp output are unaffected. That a measurement is meaningless without stating what it was measured against is true of any ratio, and it is why a published absorbance is always accompanied by its solvent and conditions.

Please log in or register to report this question.

Answer: A

A bubble in the beam scatters light out of the path, and the instrument records the missing light as absorbance. It does shorten the liquid path, but the scattering dominates and pushes the reading up rather than down. A bubble outside the beam is indeed harmless, which is why cells are filled and tapped carefully. An instrument that measures attenuation cannot distinguish absorption from scattering, and every practical precaution in this topic follows from that limitation.

Please log in or register to report this question.

Answer: B

Small differences in how a cell sits in its holder change which part of the glass the beam passes through and how it is reflected, giving a spread of readings that a single measurement hides. Settling, detector recovery and lamp output are not the reason. Repeating a measurement after disturbing the thing measured is what reveals the reproducibility of the whole procedure rather than merely the stability of the electronics.

Please log in or register to report this question.

Loading…
Tags: Spectrophotometry Fundamentals Questions and Answers || Spectrophotometry Fundamentals MCQ Questions and Answers || Spectrophotometry Fundamentals GK Questions and Answers || Spectrophotometry Fundamentals GK MCQ Questions || Spectrophotometry Fundamentals Multiple Choice Questions and Answers || Spectrophotometry Fundamentals GK || GK on Spectrophotometry Fundamentals || Analytical Instrumentation Questions and Answers || Analytical Instrumentation MCQ Questions and Answers || Analytical Instrumentation GK Questions and Answers || GK on Analytical Instrumentation