Signal Conditioning Circuits Questions and Answers

1.

The signal conditioning unit in biomedical instruments primarily performs

Exam Year(s)

Tripura Public Service Commission - 2026,

Answer: C

The signal conditioning unit performs signal amplification, filtering, and isolation.

Biological signals are tiny (ECG about 1 mV, EEG tens of microvolts) and buried in noise and mains interference. The conditioning stage between transducer and display amplifies the signal, filters out unwanted frequencies such as 50 Hz hum and baseline drift, and isolates the patient for safety, so that the signal is ready for display, recording or digitising.

Data storage, image reconstruction and waveform display are later stages that work on the conditioned signal, not conditioning itself.

Please log in or register to report this question.

2.

The primary function of an instrumentation amplifier in biomedical systems is to

Exam Year(s)

Tripura Public Service Commission - 2026,

Answer: B

The primary function is to amplify small differential signals while rejecting common-mode noise.

Biopotentials such as the ECG appear as a small difference between two electrodes, while mains interference appears almost equally on both. The classic three-op-amp instrumentation amplifier has very high input impedance, gain set by one resistor, and a high common-mode rejection ratio (often above 100 dB). It amplifies the difference and cancels what both inputs share.

Analog-to-digital conversion is done by an ADC, and impedance is measured with an impedance meter or bridge. An instrumentation amplifier is not a filter, and certainly not one only for high-frequency noise.

Please log in or register to report this question.

3.

The drift in a DC-coupled biomedical amplifier can be minimized by

Exam Year(s)

Tripura Public Service Commission - 2026,

Answer: A

Drift is minimised by using AC coupling (capacitive coupling).

In a DC-coupled amplifier, slow changes such as electrode polarisation potentials, offset voltages and temperature effects are amplified along with the signal, so the baseline wanders and the stages can saturate. A series capacitor (a high-pass filter, about 0.05 Hz for diagnostic ECG) blocks these DC and very-low-frequency components while passing the signal of interest.

Reducing input impedance would load the electrodes and distort the signal. Large input resistors add thermal noise and bias-current offsets, which make drift worse. Adding feedback capacitance shapes the high-frequency response, not the DC drift.

Please log in or register to report this question.

4.

The correct order in which data is most reliably transmitted

(i) Current loop
(ii) V to F conversion
(iii) Voltage transmission
(iv) Digital data communication

Exam Year(s)

UGC NET - 2012,

Answer: A

The key gives the order in which data are transmitted most reliably, from least to most reliable, as voltage transmission, current loop, voltage-to-frequency conversion and digital data communication, that is (iii), (i), (ii), (iv). A plain voltage signal loses accuracy with the resistance of the line and picks up noise, a current loop is not affected by line resistance, a frequency signal is not affected by amplitude changes along the line, and digital data can be checked and corrected, so it is the most reliable. The other sequences place the digital method first or last in a different order. The answer is (iii), (i), (ii), (iv).

Please log in or register to report this question.

5.

The correct sequence to be followed to build data acquisition system is by arranging the steps as follows :

(i) transducer
(ii) software development
(iii) range of parameter
(iv) signal conditioning

Exam Year(s)

UGC NET - 2012,

Answer: C

To build a data acquisition system the steps are taken in this order. The range of the parameter to be measured is first decided, step (iii), because it fixes the choice of the sensor. The transducer that converts the parameter into an electrical signal is then selected, step (i). The signal conditioning, such as amplification and filtering, is next, step (iv), and last the software is developed to read and process the data, step (ii). So the sequence is (iii), (i), (iv), (ii). The sequences that start with the software or the transducer put the choice of the device ahead of the requirement. The answer is (iii), (i), (iv), (ii).

Please log in or register to report this question.

Loading…

Other Subjects

Tags: Signal Conditioning Circuits Questions and Answers || Signal Conditioning Circuits MCQ Questions and Answers || Signal Conditioning Circuits GK Questions and Answers || Signal Conditioning Circuits GK MCQ Questions || Signal Conditioning Circuits Multiple Choice Questions and Answers || Signal Conditioning Circuits GK || GK on Signal Conditioning Circuits || Sensors and Transducers Questions and Answers || Sensors and Transducers MCQ Questions and Answers || Sensors and Transducers GK Questions and Answers || GK on Sensors and Transducers